#184

Ғылым және технология

This video demonstrates the basic process of calibrating a basic, low bandwidth analog CRT-based oscilloscope. There really isn't a substitute for having the necessary equipment and documentation to perform the job properly. The process begins with checking and adjusting the power supplies, then moves on through the CRT controls, vertical circuits, trigger circuits and horizontal circuits. In most cases, adjustments are made using small trimmers or trimpots, or small adjustable capacitors. Of course, I am not a professional metrologist, so the intent of the video is simply to give you an idea of the basic process of what's involved - as I'm sure that there are certain procedures and considerations that a professional would do that I missed. The oscilloscope used in the video is a Heath 4552 from 1989. This scope was made by Pintek, and is the same as the Pintek PS-251, which can still be purchased new (believe it or not!). Over the years, this same basic scope has been branded by the likes of Elenco, BK Precision, and many others.

Пікірлер: 269

  • @JCBudro16
    @JCBudro164 ай бұрын

    Excellent video. Glad I found it even nine years later. One thing that might be worth adding is that it's usually good to let the scope warm up for 20-30 minutes before performing the calibration.

  • @LostDeadSoul
    @LostDeadSoul6 жыл бұрын

    Nice! You gave me the courage to try to adjust my 44 year old 25MHz oscilloscope. The lines are beginning to oscillate in intensity and both X and Y axis amplitudes are off by 1cm. It was last calibrated in 30/01-85 and was due next 30/01-86 which is three months before i was born. Funny :)

  • @thats2kewl
    @thats2kewl6 жыл бұрын

    Such a good (and useful) video...I really appreciate the way you explain the steps you take. Presentation is so clear and informative. Thanks so much!!

  • @964tractorboy
    @964tractorboy6 жыл бұрын

    Very clear and instructive as always. A process I'd never previously considered to be so involved, if indeed I considered it at all. Thank you.

  • @chongli297
    @chongli297 Жыл бұрын

    Awesome video, thanks Alan! I gotta say, what a lovely scope that is. Nice sharp trace, very clean and organized controls, what a beauty!

  • @stevescnc8155
    @stevescnc81559 жыл бұрын

    Great video, very knowledgeable and concise information. Helps understanding about calibration and adjustments that can be made.

  • @samba3403
    @samba34039 жыл бұрын

    Finally a guy who talks electronics, grate set of videos that i have seen so far, keep up the good work.

  • @jonyheringer3150
    @jonyheringer31504 жыл бұрын

    This video was a great class for those looking for such important technical information! Thank you very much for recording your work; because that helped me a lot. Be able to see; how to make. Simply amazing!

  • @KX36
    @KX369 жыл бұрын

    I like this video, watching every adjustment come back in gives me the feeling that all is right with the world... :)

  • @paulmorrison1821
    @paulmorrison18214 жыл бұрын

    Great instructional video - thanks so much for taking the time and sharing your knowledge.

  • @das250250
    @das2502507 жыл бұрын

    Beautiful work , i didn't realise the steps were so many in a simple scope. Also , what a great educational process building such a scope would be .

  • @grahambambrook313
    @grahambambrook3135 жыл бұрын

    Another great tutorial video, Alan. Many thanks, it gave me the confidence to attempt a rudimentary calibration on a Tek 7403N I just managed to resuscitate and save from the scrap yard!! :-)

  • @olafwagner
    @olafwagner4 жыл бұрын

    This is a brilliant video, thank you. I just purchased a GoldStar OS-7020 scope, and I will go through exactly this process (as soon as I can track down an elusive service manual)

  • @Xanthyn
    @Xanthyn5 жыл бұрын

    Even though I have a different scope, this was a helpful intro to help me understand what I'm looking for when calibrating my own.

  • @MikeNHOC
    @MikeNHOC9 жыл бұрын

    That was very informative. Thank you for creating the video.

  • @321reh
    @321reh9 жыл бұрын

    Thanks Alan for another Great Video!!

  • @bobjacko5658
    @bobjacko56587 жыл бұрын

    Enjoyed your calibrate video and it pointed out how I do, indeed, have to find the instruction manual for my NRI model 2500 scope. Many of your other videos, however, are also helping me out to learn how to operate the scope.

  • @billbates7810
    @billbates7810 Жыл бұрын

    Really nice video, very well put together!

  • @paultopolski6595
    @paultopolski65959 жыл бұрын

    Alan, Superb video as always. This is a perfect warm up for calibrating a couple of Heathkit scopes, an IO-4510 and IO-4550 as well as an Elenco 1325. All together, I spent $65 for the three of them a various ham fests. These scopes will be good practice because four Tek 7000 mainframes and about twenty plug-ins will follow. Thanks for the great work. Thanks, Paul, W1SEX

  • @88merc300e
    @88merc300e9 жыл бұрын

    Very good info. Why would anyone dislike this? Thanks Alan.

  • @judges69
    @judges697 жыл бұрын

    Thank you for this video, its just a hobby for me and you helped a lot.

  • @Pwaak
    @Pwaak9 жыл бұрын

    Excellent lesson all around! Looks like it was fun to kick back and tweak "ol' betsy"! Actually, this procedure is up next after a Eico 460 rebuild!

  • @BM-jy6cb
    @BM-jy6cb2 жыл бұрын

    W2aew - your content is top notch. I really don't understand why the KZread algorithms constantly push celebrity court cases or how much can you eat challenges when this sort of subject matter is all I ever watch!

  • @magdalenoarellano1841
    @magdalenoarellano18419 жыл бұрын

    otro gran video de mister muy instructivo y motivacional muchas gracias

  • @Yazeed834
    @Yazeed8346 жыл бұрын

    A great job as usual!

  • @pa4tim
    @pa4tim8 жыл бұрын

    Nice information, like allways, with sound advise, brings me back to my first serious scope cal, a Tek 465. I got it right but it took a lot off time (like 15 hours or so) and I learned a lot from it, and for most that you first have to check if you understand every step and have all the gear needed. Missing one instrument can screw up the whole operation. I really learned that when I calibrated a 1ns pulse generator with a 150 MHz scope. It looked perfect until I got a 1GHz scope and had to do it all over again. You often need better testgear as the gear you are adjusting or repairing. Some tips for your viewers if you do not mind: Do not think to light about calibration and adjustments of test gear, it is very easy to make it (much) worse. A tip I learned long ago from an experienced meterologist, is to mark the settings of adjustable components before you adjust them, so that if you screw up, you can get it back to its old state. A picture will help too. What I sometimes do is making a few high res pictures and marking all adjustment points with their name/number and value and in more complex steps the number/paragraph that step has in the manual. That can speed up things a lot and forces you to read and understand every step in the process before you start. Do not do partial calibrations unless you really know what you are doing and not because you do not understand a step in the process. It is very easy to make it worse. Back then it was a hobby (it still is) . Nowadays I make my living from repairing test-instruments and calibration gear. I do no calibrations (not my idea of fun ;-) ) but depending on the work I did, I sometimes have to calibrate it, or at least check it after repair.

  • @w2aew

    @w2aew

    8 жыл бұрын

    +pa4tim Good tips - thank you!

  • @SIXSTRING63
    @SIXSTRING635 жыл бұрын

    I have the same scope branded with the Elenco name. Came with my NRI home study course in the early 90’s. Still use it. Works great. Maybe I’ll go through it and touch it up a bit.

  • @jorgesantos1225
    @jorgesantos12259 жыл бұрын

    Great video! Thank you.

  • @blaaaaaaarghable
    @blaaaaaaarghable9 жыл бұрын

    You forgot to mention maintaining the correct tongue angle when making adjustments :p

  • @w2aew

    @w2aew

    9 жыл бұрын

    Oh, well that goes without saying! Why do you think you never really see me on camera!

  • @dave-d

    @dave-d

    7 жыл бұрын

    There's a channel for that!

  • @boonedockjourneyman7979

    @boonedockjourneyman7979

    6 жыл бұрын

    The only person I know who has the angle problem has a whole ton more in other areas.

  • @sbybill3271

    @sbybill3271

    4 жыл бұрын

    :)

  • @seancastledine8983
    @seancastledine89835 жыл бұрын

    Good video. Nice work.

  • @THESKYMASTER
    @THESKYMASTER9 жыл бұрын

    Very good video, thanks!

  • @dave-d
    @dave-d7 жыл бұрын

    Thank you. I am in the midst of an HP Agilent 1740A mega repair / refurb and have all this to come. Still chasing vertical jitter in this one. The gremlin has hidden itself well in this one!

  • @gamingSlasher
    @gamingSlasher9 жыл бұрын

    Top notch! All my thumbs up!

  • @clems6989
    @clems69898 жыл бұрын

    the first job I had on a bench, was equipped with one of these scopes. I realize they were rebranded under different names sometimes. But my goodness just Imagine how many of these scopes were sold to ametuers, hobbyist, schools etc. And how many good techs were self taught using these. Back in the day a 20 or 25 Mhz dual trace scope was the entry level scope, And extremely useful even to this day..I really enjoyed watching you calibrate this old girl...Sorta like watching someone do a tune up on a 57 Belair. .......lol....N8QZH

  • @w2aew

    @w2aew

    8 жыл бұрын

    +Matthew Miller Yes, I've seen this same scope with at least 4 different nameplates on it. A nice, very useful, basic scope.

  • @davecc0000
    @davecc00009 жыл бұрын

    Another great vid Alan. Many of have heard of cal'ing a scope; I for one didn't know how involved it was... Thanks!

  • @igsaturation
    @igsaturation9 жыл бұрын

    Nice work. That Heath scope looks almost new. Aside, the quality of the documentation are pretty rare to find in modern gear. It does serve to warn others looking to get an analog scope of unknown status over eBay that getting the parts replaced in one thing, adjusting it to factory performance is another.

  • @Gw0wvl
    @Gw0wvl7 жыл бұрын

    Another Great Video ... Thankyou .

  • @pji6494
    @pji64949 жыл бұрын

    Job well done!

  • @Valenorious
    @Valenorious9 жыл бұрын

    Probably important to emphasize is that it needs time to warm up and settle before you make calibration adjustments. But that goes for use as well.

  • @vhm14u2c
    @vhm14u2c6 жыл бұрын

    The 60mhz scope I just bought used is a bk precision 2160a, has almost identical controls on front panel, including component test on bottom row. Bk did not provide much for calibration in their manual. I have found that hk so-4552 manual with the specific adjustments you done. I’m not going to do any, but may check references from the manual if I find problems down the road. Thanks for sharing!

  • @gixerags750
    @gixerags750 Жыл бұрын

    I have watched this over 20 times...such a brilliant informative how to Most 20 -25mhz CRT oscilloscopes look they are the Pintek chassis, I have seen, my Telinstruments, Sony, even the Hitachi V20, look the same, they all seem to havea common REAR panel , connectors The Heath 4552 service manual is easily available on the Web, unfortunately my Telinstruments TI5252 was not.

  • @gixerags750

    @gixerags750

    Жыл бұрын

    Hi Alan, would there be any reason to or not to replace the Electrolytic capacitors before this calibration procedure,? It's just I plan on following this video and would like to replace caps while it open. Sorry to be a pain Cheers

  • @pcrengnr1
    @pcrengnr12 жыл бұрын

    Alan thx for sharing. As for the transient response adjustment, the picture in the manual shows ringing and you even commented about the ringing. I like what you ended up with, a nice square wave with no ringing. I'm with you, if I could've adjusted it so that I didn't get ringing but got the squarest wave I could I would've too. I wonder why they suggested adjusting the transient response so that ringing was present. By the book a little bit of ringing does give the widest bandwidth before falloff like a Chebyshev filter. However, the downside is the time domain suffers. I still like your solution. Again, thx for sharing.

  • @frecuenciajuaniyo
    @frecuenciajuaniyo9 жыл бұрын

    very nice video

  • @mdzacharias
    @mdzacharias9 жыл бұрын

    Just a thought - since I work on stereo equipment for a living, I think I would do the ch. 2 vertical sensitivity adjustment by overlaying it over ch.1 (after it has been adjusted) and adjusting so that the two channels match as closely as possible. Just my .02...

  • @lwilton

    @lwilton

    2 жыл бұрын

    I've tried that, and it doesn't work as well as you would think. The problem is the trace lines are wider than the gradicule lines, and they are the same color, so you usually end up with the second channel being worse calibrated than the first channel. It is easier to do both per book, then A/B compare the results with the same signal fed to both channels, and make any tweaks that might be required to make them exactly match, if they will. Since it is an analog machine, they won't always match perfectly.

  • @millerellen2258
    @millerellen22586 жыл бұрын

    Thanks for you video

  • @anthonyroby296
    @anthonyroby2969 жыл бұрын

    Very tinely since I just picked up a Tek 2225. I haven't played around much with it yet to see if it's in need of calibrating.

  • @brikaf6001
    @brikaf6001 Жыл бұрын

    Excellent

  • @MrMac5150
    @MrMac51509 жыл бұрын

    Very good.

  • @jimis04
    @jimis044 жыл бұрын

    Congrats! And liked! I have the same exact oscilloscope but as a brand it writes MiteK mk4300..im so happy that the only complete oscilloscope calibration on KZread is on my oscilloscope... Must say that the insides look completely the same as well. just the operation led is red on mine, not green but when possible I will change it because it's annoying and looks awkward

  • @boonedockjourneyman7979
    @boonedockjourneyman79796 жыл бұрын

    Excellent. Thank you. Why are “experts” unwilling to try things like this?

  • @dri50
    @dri505 жыл бұрын

    I just bought a Heath IO-4550 for hobby use and audio work. It's not coming with manuals, had to buy the assembly and user manual separately. I would think a quick inspection of the power supply electrolytics will be my first step before attempting calibration. Thanks for the video.

  • @barn5923
    @barn59237 жыл бұрын

    thanks enjoyed this very much

  • @Wil_Bloodworth
    @Wil_Bloodworth3 ай бұрын

    This is excellent! I wish I could find the equivalent calibration procedure for my Sencore SC3100. I'm not sure those instructions exist.

  • @joeobrien1888
    @joeobrien18888 жыл бұрын

    Alan, this is FANTASTIC. About the best I've seen in the way of calibration, you sound like you might be a teacher. I worked with these scopes for about 25 years (DEC) and have the exact one you use in this video. Do you know where I can get the Service Manual? I've looked, but it's pretty hard to find. Thank you again for such a good and professional video.

  • @w2aew

    @w2aew

    8 жыл бұрын

    +Joe O'Brien I'm not sure where to find the manual. I happen to have gotten one with the scope.

  • @yurin5227
    @yurin52279 жыл бұрын

    Hi, I like your video, very easy to follow. I've got a bit of a problem, wondering if you could help. I picked up the same scope off an auction, but there's an electrical problem I can't quite track down. Is it possible to get a scan of the electrical diagram that came with it? Mine came with no manuals. I'd really appreciate it.

  • @tanukicyber
    @tanukicyber9 жыл бұрын

    Thanks for the informative video. What did you use as you signal sources? I have a Tektronix 465, like you too. I wish you had done that scope, but I'm not complaining! It's nice to see what you had to do, so I'm aware of what is coming down the road for me and my oscilloscopes.

  • @w2aew

    @w2aew

    9 жыл бұрын

    tanukicyber I used a Tek AFG3252 - fast and accurate enough for these old analog scopes.

  • @BruceNitroxpro

    @BruceNitroxpro

    6 жыл бұрын

    make sure you recap it first... and test the power supplies... making sure you use 117 vac through a good isolation transformer at test time

  • @rostlaur
    @rostlaur9 жыл бұрын

    If your videos get the merit that they deserve I would suggest that your introduction salute the viewers from the 22nd century and after ;^)

  • @w2aew

    @w2aew

    9 жыл бұрын

    Thanks! But really wonder how much of this will be relevant in the next century...

  • @caulktel
    @caulktel6 жыл бұрын

    Good video Alan, you make it look easy. Do you know where I might find service manuals for my two Kikusui oscilloscopes. I have a COS5040 and a COS5100. They both work great, but could stand to be calibrated.

  • @w2aew

    @w2aew

    6 жыл бұрын

    You might try the manufacturer - they will sometimes have the service manuals available for free. You may have to contact their customer service.

  • @MikeRR142
    @MikeRR1425 жыл бұрын

    Thanks for this great video! Very helpful. Was wondering about the 50ohm terminator, why was that used? Didn’t see where that was required in the video shots of the manual you were using. Is it because you are not using a probe but directly connecting to the waveform generator?

  • @w2aew

    @w2aew

    5 жыл бұрын

    I used it to minimize reflections and distortion from my signal generator, especially on the square wave signal.

  • @larryrenaud7462
    @larryrenaud74629 жыл бұрын

    Great video, always looking forward to your next one. 73 VE3BHM Larry

  • @FullElectronic
    @FullElectronic8 жыл бұрын

    I just need to know what nomenclature HP uses for trigger offset.

  • @acolisys
    @acolisys3 жыл бұрын

    Another real master class, Alan ! By the way, I'd like to thank you for the spectacular knowledge that you spread through your videos ! Do you have any reference material / video about Delay Time / Dual Time Base oscilloscope calibration, please ? Merry Chrismas & Happy New Year ! Wish you all success, heath, peace, happiness and the best greetings from Brazil !... ^_^

  • @w2aew

    @w2aew

    3 жыл бұрын

    This video should be what you need: kzread.info/dash/bejne/Z2im0dxxebjVhqw.html

  • @acolisys

    @acolisys

    3 жыл бұрын

    @@w2aew Excellent, I'll watch it in a moment. Thank you ever so much !...

  • @mikezike1
    @mikezike14 жыл бұрын

    Thanks for the wonderful video. Although I have both digital and analog scopes, I really like the analog for the real time stuff. Is there an alternative to the Time Mark generator for calibration?

  • @w2aew

    @w2aew

    4 жыл бұрын

    Any signal or function generator whose frequency accuracy is better than the scope you're calibrating, you can use.

  • @GaryGrove
    @GaryGrove Жыл бұрын

    Enjoyed your video Alan! It does look strange to see you calibrating a unit that is less than 1 Gig!! 73's , Gary Grove

  • @thecow2756
    @thecow27564 жыл бұрын

    this scope has been also sold as teklab osc-25 and mine has a issue where the up/down position is faulty on both channels and has to be set fully to one end to show on the screen so need to do checks on where the fault is at

  • @thrill4kill143
    @thrill4kill1436 жыл бұрын

    What signal source did u used to "cal" bandwidth, response time and time base?

  • @mian2zi3
    @mian2zi39 жыл бұрын

    Great video, thanks! I just picked up a used Tek analog scope I'm going to try to repair/calibrate. This is my first time working with with a device with high CRT voltages. Where did you get that insulated screwdriver? I'm having trouble finding something spec'ed above 1kV, let alone the voltages in the scope (14kV).

  • @w2aew

    @w2aew

    9 жыл бұрын

    mian2zi3 The screwdriver was made completely of plastic - it was purchased as part of an "alignmnent tool kit", typically used to adjust the RF coils in radio and TV receivers.

  • @eldenchang
    @eldenchang9 жыл бұрын

    Hello Alan, Thanks for another great video. Where can I get these small and nice BNC cables you are using? Are they sold like this or did you make them yourself? What kind of cable is it? It is so nice and thin. Thanks

  • @w2aew

    @w2aew

    9 жыл бұрын

    Those are made using RG-174 type coax. I have made some myself, but I have also purchased several from eBay. I've had good luck with items from this ebay vendor: stores.ebay.com/adapterconnectorpigtailcable/

  • @eldenchang

    @eldenchang

    9 жыл бұрын

    ***** Thanks for the advise of the cables, I ordered some from this vendor an hope they are good quality.

  • @tubical71
    @tubical719 жыл бұрын

    Thanx for this one, as i now know that i do not have to open my "new" "Vuko VKS220-16" scope to rotate the CRT for an all flat trace, as the little trimpot, you´re revering to at 06:40 has only "TR" written on the silkscreen of that vuko, and it´s just next to intensity, therefore i thought this is to trim the intensity. Now i know better. As that "TR" is AKA trace rotation. Thanx a lot for this, rather tiny but very useful, hint, as it prevents me to work that scope CRT :)

  • @johnnaggo4002

    @johnnaggo4002

    9 жыл бұрын

    I have the same scope. For the analog part I can use the appropriate Hameg manual, but do you have any info on the digitizing section?

  • @tubical71

    @tubical71

    9 жыл бұрын

    I got the whole documentation from vuko, a big solid A4 letter folder, about 300+ pages....

  • @krisztianandre

    @krisztianandre

    9 жыл бұрын

    TubiCal do you know how to display the stored waveforms, or you can only access the memory trough the gpib interface?

  • @tucker934
    @tucker9342 жыл бұрын

    Just gotten a Hitachi V-209 20MHz, and I can’t seem to find the calibration manual/instructions for it. The V-212 is also A 20MHz scope and there’s one for it, but coming up dry on the V-209 any help?

  • @gregorystevens5173
    @gregorystevens51738 жыл бұрын

    Hi Alan, thanks for the video and the very clear explanations. I'm somewhat new to the O-scope world and certainly new to scope calibration. I would love to have a generator like a 3252 but it's out of my budget at the moment. Are there less expensive (

  • @w2aew

    @w2aew

    8 жыл бұрын

    +Steve Novosel Some of the new generation of inexpensive DDS based signal and function generators would be sufficient for many of the needs of checking/adjusting an old analog scope.

  • @stephennovosel

    @stephennovosel

    8 жыл бұрын

    How critical is it to find one that includes a pulse generator with 1 ns or less rise time (for purposes of calibration of a 25MHz Heath 4552)?

  • @w2aew

    @w2aew

    8 жыл бұрын

    +Stephen Gregory Not super critical given that the scope's risetime is about 14ns. But, I'd want to have a pulse that is a few ns or so. This can easily be done with some fast logic chips. See my video on the Schmitt Trigger oscillator. Note that several viewers have made some boards available.

  • @aposhanski
    @aposhanski5 жыл бұрын

    Hello Alan. My name is Alexander R2DHL. I have got an old Tek 2445b and this is my favourite scope. About a month ago an error appeared: test 4-11 (calibration constant checksum) and everything pointed on a nvram chip NEC D4364C-12LL fault. So I have replaced it with the UT6264CPCL-70LL chip (it has more memory and a litle bit faster). The battery was made in 1988 and there was 3.6 volts on it but I have made a decision to replace it too. Now I need to calibrate my scope. The calibration manual that I have found on internet is not easy to understand because I am doing that for the first time. 2445b is a pretty common scope but I could not find any video instructions for the calibration procedure. I will be very appreciated if you will help me with my problem.

  • @FullElectronic
    @FullElectronic8 жыл бұрын

    Hi Alan. What is the name of the adjust made at 16:02 where you adjust the wave to zero on the left side of screen? I have a hp1740A in need of this adjustment. My listening in english isn't so good but, you said trigger offset?

  • @w2aew

    @w2aew

    8 жыл бұрын

    +CYBERRD58 Yes, it is the trigger offset adjustment - it is adjusted with the trigger on AC coupling, trigger level control set to the midpoint, then adjust the trimmer (assuming there is one) so that the signal begins at the middle.

  • @FullElectronic

    @FullElectronic

    8 жыл бұрын

    Thank you.

  • @FullElectronic

    @FullElectronic

    8 жыл бұрын

    +w2aew Thank you.

  • @gixerags750
    @gixerags750 Жыл бұрын

    Hi , i could really use your advise, my channel 1 is receiving a 1khz signal but my vertical position control will not bring the trace high enough to centre the trace.. channel 2 works fine.. it is a a Telinstruments TI 5252 25MHZ scope, . The correct waveform is there but is hanging way below the bottom of the scrren..very frustrating. Cheers

  • @larryrenaud7462
    @larryrenaud74629 жыл бұрын

    Great video just picked up this scope, question where did you find the service manual.

  • @w2aew

    @w2aew

    9 жыл бұрын

    The service info is in the back of the users manual that came with the scope.

  • @larryrenaud7462

    @larryrenaud7462

    9 жыл бұрын

    I got the scope but no service manual, can't find one on line 73VE3BHM

  • @scottson2
    @scottson28 жыл бұрын

    Very informative video.. What brand and model of HV robe is that? I'm looking for one currently.. thanks.

  • @w2aew

    @w2aew

    8 жыл бұрын

    +scottson2 It's an old RCA WG-297 - it can be seen in more detail in my video on High Voltage Probes: kzread.info/dash/bejne/X5OArLqEfMqvicY.html

  • @Giovanni2862
    @Giovanni28624 жыл бұрын

    Could you help me, what instruments are needs to calibrate a Tektronix 2465 ?

  • @TheRogerx3
    @TheRogerx39 жыл бұрын

    Thank you for yet again for an interesting video.. but sadly I wont under take doing mine - High voltage calibration is not for me.. good advice noted.

  • @gixerags750
    @gixerags7502 жыл бұрын

    Love your channel..is there any way of making this procedure less dangerous? eg an isolation transformer etc? or are the CRO votages , unavoidable just way too high? Thanks in advance.

  • @w2aew

    @w2aew

    2 жыл бұрын

    The CRT voltage is very high, not much you can do other than use extreme caution.

  • @gixerags750

    @gixerags750

    Жыл бұрын

    ​@w2aew can this calibration process still be partially successful by skipping the high voltage crt step? 73

  • @MrDIMITRIOUifyable
    @MrDIMITRIOUifyable8 жыл бұрын

    i did find your video very interested , . i do have an oscilloscope in my position the DM64 Analogue storage oscilloscope i have not a clue how to get it working for so long . i do have all manual seen your video i am going to have a go

  • @kgsalvage6306
    @kgsalvage63066 жыл бұрын

    Hello , Been watching your videos on oscilloscopes. I just got a Sperry 620C . It's in excellent condition , but I have an issue with the vertical position . It won't come down to center on either channel . It's all towards the top and off the screen . Any thoughts on what to look at . Any help would be greatly appreciated . Thanks

  • @w2aew

    @w2aew

    6 жыл бұрын

    I'm not familiar with that scope, but if the vertical offset problem is common to both channels, then you can rule out the vertical preamps. I would focus on the CRT vertical plate amplifier and driver stage.

  • @bigbread9000000
    @bigbread90000009 жыл бұрын

    I have the same scope with the pintek brand name. I was wondering if there was a calibration for the cal lug?

  • @w2aew

    @w2aew

    9 жыл бұрын

    Yes, there are adjustments for frequency, duty cycle and amplitude. I didn't include them in the video because I didn't want to give the impression that the CAL signal could be used as a cal reference. It's only intended purpose is for probe compensation. Even if it is adjusted properly, it won't be stable enough to be a cal reference signal.

  • @Kennynva
    @Kennynva7 жыл бұрын

    I have the problem at 18:51 in the video..where the vertical pips are not in alignment the further you get from left to right...I can find no adjustment in this old scope (Hitachi V-302)...what is the pot for this adjustment called in your scope..Thanks for all the fine work you do...I can tho turn the SWP VAR to get them to line up...

  • @w2aew

    @w2aew

    7 жыл бұрын

    It looks like (from the schematic in the manual on Electrotanya) that the calibration adjustment for sweep speed might likely be a variable capacitor called CV510 shown on page 26.

  • @Kennynva

    @Kennynva

    7 жыл бұрын

    Strangely enough it was a pot not a VC and it was called (X1 H GAIN). I hate moving things to find out which one to adjust but sometimes you have to take the chance..Im going by a V-202 calibration..and seems close..just different VR and CV numbers...

  • @JIMO415
    @JIMO4154 жыл бұрын

    Hello sir, I have watched a few of your scope videos and they are very helpful. I appreciate them. I have been playing with Tektronix 465 units and now have 3 of them. I was able to get the first 2 going and I have a question as I dig into the third regarding the use of an old Triplett 630 VOM to measure the HV supply. It has a 6000 volt range but I was wondering if it has too low of an input impedance to measure the -2450 on the Tek 465 (doing this measurement hands off of course). I understand the HV probe would be the better and safer thing to use here ( I do not have). I was just wondering your thoughts if you care to share. Thank you , Jim

  • @w2aew

    @w2aew

    4 жыл бұрын

    It should be OK. THe 630 is a 20,000 ohms/volt meter, so it will have about 120Mohms of resistance when used on the 6000 volt setting. Just remember to use the special 6000V jack on the front panel, and be EXTREMELY cautious as this HV can be lethal. Make sure the probes you are using have insulation rated high enough.

  • @JIMO415

    @JIMO415

    4 жыл бұрын

    @@w2aew Yes and thank you sir. yes, probes are a concern at this point. Jim

  • @JIMO415

    @JIMO415

    4 жыл бұрын

    Yes sir it worked well, thank you. I was able to confirm complete proper voltage of all supplies. I now just have to find out why there are no traces on either channel and just a dot when beam find is pressed.

  • @w2aew

    @w2aew

    4 жыл бұрын

    @@JIMO415 If you can move the dot back and forth with the horizontal position control, then that tells you that the horizontal driver (and likely horizontal amp) are OK, so then focus on the sweep generator circuit. Nice thing is that you have 3 of them - one to test with, one to verify what the waveforms and voltages are supposed to look like, and one to troubleshoot - ideal situation!

  • @JIMO415

    @JIMO415

    4 жыл бұрын

    @@w2aew yes, I agree at first. As I get into it, I can't duplicate the waveforms in the manual with with what I find on the test points of the working scope. But it was late and I was tired so ? This one will be a learning struggle! Yes, the dot does move horizontal back and forth about an inch as well as the vertical up and down with beam find pushed in. It appears as if each channel has it own sweep generator. So if there is no trace in either channel, it must be a single, common to each channel issue I would think?? we will see eventually I hope. Jim

  • @bigbread9000000
    @bigbread90000006 жыл бұрын

    I have the Pintek brand of basically the same scope with delay sweep. I have a slightly bent trace from left to right of screen(not the rotation), I think I caused it by not having my probe set at 10x, when I was measuring a higher voltage. I was wondering, is there any adjustment for this or did I damage the plates in the tube?

  • @w2aew

    @w2aew

    6 жыл бұрын

    I'm not sure I understand your description of the distortion. But, I will say that it is highly unlikely that you damaged anything - certainly not the plates in the tube unless the scope was dropped.

  • @bigbread9000000

    @bigbread9000000

    6 жыл бұрын

    Maybe a better description would be to say that when you put the coupling lever in any position the trace has a slight arc(like a rainbow) just ever so slightly though!

  • @w2aew

    @w2aew

    6 жыл бұрын

    Hmmm - I wonder if the chassis is slightly magnetized.

  • @larryrenaud7462
    @larryrenaud74628 жыл бұрын

    Hi Alan, again I need your advice. The scope I was restoring looks great using your suggestions, of Windex and alcohol. The potentiometers need cleaning. The scope is a Tektronix 422, some sites say don't use contact cleaner but replace them, that would be cost prohibitive, what is your suggestion. Thanks Larry

  • @w2aew

    @w2aew

    8 жыл бұрын

    I've always had good luck with careful use of DeOxit.

  • @mrnmrn1

    @mrnmrn1

    8 жыл бұрын

    +w2aew Hi, I'm new on your channel. I'm not familiar with DeOxit, is it leaves some oil residues, or completely dries out? I have always afraid using oily contact cleaners in oscilloscopes, than later my fears has been confirmed when I saw a busted scope, which had its Y attenuator switches insulted with oily contact cleaner. The oil penetrated inside the compensating trimmer caps, made it impossible to set them properly, as the oil changed their capacitance so much, and probably increased their dissipation factor at high frequencies... Even the probe compensating 1kHz square wave of the built-in generator looks horrible in some ranges, while other ranges are better, but non of them are perfect. In your opinion, all of the trimcaps needs to be replaced in the attenuators, or maybe an ultrasonic cleaner filled with isopropanol would help? This isn't my scope, I've seen it at work, but it's just so sad to see that beautiful 100MHz Iwatsu in that condition... I personally have a Leader LBO-508, a HP 1223A, and the Hungarian clone of the Tektronix 647 (EMG 1555 / TR-4653), all of them needs to be calibrated, but I have no calibrated generators, so I will probably do the calibration via comparing the 'scopes under test' to a calibrated scope at my workplace.

  • @w2aew

    @w2aew

    8 жыл бұрын

    +mrnmrn1 Yes, you do have to be careful with these cleaners that leave a lubricant. Use them sparingly, avoid overspray, etc. In cases where you have open contacts, using ordinary white paper soaked in isopropyl alchohol used as a wipe on the contacts works well. It is hard to say about the trimcaps. Replacement would likely be best, but tedious and expensive, so it may be worth trying some other methods first.

  • @mrnmrn1

    @mrnmrn1

    8 жыл бұрын

    +w2aew Thank you for your answer. I used to use the white paper trick on open contacts, but usually I start with dry paper, as it rubs the surface more than isopropanol wetted paper. My second step used to be the isopropanol wetted paper, to remove the rubbed-off oxide residues. If the oxidation is more serious, I used to wet the paper with one of those ugly, oily contact cleaners, as they are great silver oxide solvents. This is the only way I use lubricant type contact cleaners in sensitive equipment. I used to use thick silicone grease to lubricate contacts (especially in potentiometers), as it lasts longer, and doesn't penetrates everything on the board, just stays on the contacts for years, protecting them from oxidation.

  • @josealfredobh
    @josealfredobh4 жыл бұрын

    Hi, is there any procedure for checking the calibration status of the analog oscilloscope, without disassembling it ?

  • @w2aew

    @w2aew

    4 жыл бұрын

    You basically have to run through a performance verification, which is putting known signals into the scope and observing if the scope is displaying their properties properly.

  • @JoaoFerreira-bi8wt
    @JoaoFerreira-bi8wt9 жыл бұрын

    Is the ripple at 11:38 normal, when you use the vertical magnification? I also get it on a second-hand analog oscilloscope that I bought.

  • @w2aew

    @w2aew

    9 жыл бұрын

    For the most part, yes. This is a very sensitive setting, and the scope is picking all sorts of noise because the case is off of it.

  • @JoaoFerreira-bi8wt

    @JoaoFerreira-bi8wt

    9 жыл бұрын

    ***** Thanks! I thought it might be out of calibration.

  • @nhnchannel5426
    @nhnchannel54265 жыл бұрын

    Can you share this document? thank you very much

  • @rainbowcomputers212
    @rainbowcomputers2125 жыл бұрын

    thx

  • @worroSfOretsevraH
    @worroSfOretsevraH6 жыл бұрын

    Hi Alan. I have a 20MHz GW-Instek GOS-620 scope, but unable to find a calibration manual for it. I saw on Tek's website that they perform such calibration, so I assume they have the manuals. I was wondering if maybe you have one, or can get one somehow? Would be of great help, so I can finally calibrate my oscilloscope. Thank you.

  • @w2aew

    @w2aew

    6 жыл бұрын

    I"m sorry, but I don't have access to service department documentation. From a teardown video I saw on KZread, this scope looks to be easily servicable, and the trimmers on the two boards look to be clearly labeled. It shouldn't be too hard to figure out a bulk of the calibration steps.

  • @worroSfOretsevraH

    @worroSfOretsevraH

    6 жыл бұрын

    Yes you are right. I've already done half of it, but still having problems with the add-balance adjustment. No matter how I set it, there is always some error when I switch to ADD. Thank you.

  • @w2aew

    @w2aew

    6 жыл бұрын

    I don't know about you, but I very rarely (if ever) use the ADD mode except for the very occasional use of ADD INVERT to get a measurement across a component. So, I wouldn't sweat it too much. I saw that Elektrotanya has the schematics for this scope, but no calibration instructions. Might be worth writing GW-Instek through their site to see if they can provide this for you.

  • @w2aew

    @w2aew

    6 жыл бұрын

    Here is the calibration adjustment instructions for that scope. drive.google.com/open?id=1YHxSVJvriqylNcCsGmSWlb5qIdEffw2x

  • @worroSfOretsevraH

    @worroSfOretsevraH

    6 жыл бұрын

    Wow! Very kind of you. I was trying to contact the manufacturer, but received no response . This world would be a much better place with more people like you. 73.

  • @HA7DN
    @HA7DN4 жыл бұрын

    I'm jealous. I am in the middle of repairing & calibrating my EMG-4656, 25MHz Hungarian scope (tek copy I think) from the end of the 1970's. Let me tell you, your is a hundred times more well built than this piece of mess. Lots of wires, connectors, most of the board is hard to access, darn hard to disassemble... Fortunately, I have a digital copy of the original manual, including the full schematics and calibration instructions.Sadly, the latter are very inaccurate, especially on how to locate the adjustment controls - referring to the wrong picture, etc. Sometimes it's straight wrong, as it mentions to adjust pots in a compleatly unrelated circuit, which will be adjusted again in the relevant section... Now I spent a week or so writing down every instruction again with the actual positions of the controls, sometimes traced back from schematic or board layouts, and just figured that I somehow broke the HVPSU while searching for the controls in there. No discharge resistors on any of the caps, so I've put the whole thing aside (also, I'll have an exam in a week), before I dare to open it again... I also lack some of the necessary equipment to properly check everything, but I have most of the stuff needed to actually adjust things. I won't ever touch anything in the HVPSU while it's on, for example. I am still pretty sure I can do with the stuff I have, as seeing the mechanical build, this is one of the pieces of eastern tech that is a miracle to work, but unbreakable.

  • @Kennynva
    @Kennynva7 жыл бұрын

    At 9:35..which pot, or should I say what is the adjustments pot name being made to make it stay the same when a and b are in the ADD mode..???

  • @w2aew

    @w2aew

    7 жыл бұрын

    According to the manual, that would be VR343.

  • @Kennynva

    @Kennynva

    7 жыл бұрын

    i meant on any scope what is the adjustment name...I am the one email you about the manual on the scope Hitachi V-302. Like one adjustment is the gain...another is the astigmatism...which name of the adjustment, adjust the difference when on ADD, or DUAL..thanks for your help..

  • @carlapeterchris
    @carlapeterchris9 жыл бұрын

    I have the exzact same Heath 4552 oscilloscope. Where would I get the service manual for it?

  • @w2aew

    @w2aew

    9 жыл бұрын

    I'm not sure - the user manual (with the calibration info) came with mine.

  • @allprogear
    @allprogear6 жыл бұрын

    My scope display is not blue on my Heath 4552?What seems to be the problem?

  • @w2aew

    @w2aew

    6 жыл бұрын

    My scope has a blue filter on the face of the CRT. It could be that yours is missing.

  • @jimkirk360
    @jimkirk3606 жыл бұрын

    I'm looking for the manual for a leader lbo-514a if anyone has any info please let me know. Thank you

  • @Roy_Tellason
    @Roy_Tellason3 жыл бұрын

    Speaking of having the proper documentation, can you point me to where I can get a service manual for my Tek 2246?

  • @w2aew

    @w2aew

    3 жыл бұрын

    Here are two sites, one is actually on tek.com: www.tek.com/manual/2246-service-manual w140.com/tekwiki/images/c/c6/2246_service.pdf

  • @Nox665
    @Nox6654 жыл бұрын

    18:30 what is that "time mark" on the screen? Just a short pulse?

  • @w2aew

    @w2aew

    4 жыл бұрын

    It's an "exponential fall" waveform available on many arbitrary function generators. This is very similar the waveform used on many legacy time-mark generators. Any periodic waveform (square, etc.) can be used as long as it's frequency accuracy is sufficiently better than the scope's timebase accuracy.

  • @Giovanni2862
    @Giovanni28626 жыл бұрын

    Could you explain how to calibrate the Tektronix 2465?

  • @w2aew

    @w2aew

    6 жыл бұрын

    This is much more involved than the simple analog scope shown in this video. The calibration procedure is outlined in about 30 pages of the service manual, which you can find here: w140.com/tek_2465a_2467_service.pdf

  • @Giovanni2862

    @Giovanni2862

    6 жыл бұрын

    w2aew I have download this www.download-service-manuals.com/download.php?file=Tektronix-1780.pdf&SID=4nqic6fff8m1u805ml264m81l5

  • @Giovanni2862

    @Giovanni2862

    6 жыл бұрын

    w2aew It is need a Time Marker Generator but I don’t have it.

  • @TheHarpanOnly
    @TheHarpanOnly4 жыл бұрын

    Do you have digital copy of that scope? I have an old oscilloscope, not same brand, but has very similar button layout.

  • @w2aew

    @w2aew

    4 жыл бұрын

    Here is the manual I used: drive.google.com/file/d/0B8Kvsw95jCqITG9XcDc0QnAwbHM/view?usp=sharing

  • @TheHarpanOnly

    @TheHarpanOnly

    4 жыл бұрын

    @@w2aew thank you for your kindly reply and support. Hope the best for you. DE YD2BLH

  • @TheHarpanOnly

    @TheHarpanOnly

    4 жыл бұрын

    Your schematic is 99% identical with my oscilloscope. And I successfully repair it becausd that schematic. I just did it. Laat night I want to calibrate my oscilloscope, and I am shocked to realize that the pcb layout is 100% different. :D

Келесі