Cheap PH Meter won't calibrate? This is how to fix it. If this doesn't fix it then nothing will.

Пікірлер: 72

  • @yclare1
    @yclare1 Жыл бұрын

    Thanks so much this video is so informative. Glad I watched before calibrating my new Ph meter.

  • @flexiblebirdchannel
    @flexiblebirdchannel Жыл бұрын

    Yes, these pH probes dry out if stored in their cover, so soak before use for 24h preferably in 3m KCl solution but tap water will do also as stated in the manual. The problem is the calibration, it does not work as mentioned in the manual. The meter will adjust to 4.01 if stored in anything that gives a readout below 6.5 and cal is pressed longer than 3s, will calibrate to 6.86 if stored in anything that gives a readout between 6.5 and 7.5 and to 9.18 if stored in anything that gives a readout above 7.5, so the region where it goes to 6.86 is very narrow. And it takes some time to stabilize, give it a minute in buffer solution. By removing the battery, it resets jammed calibration. The extensive cleaning is usually not neccessary, only milk should be cleaned with pepsine. The problem I have is the 6.86 buffer solution, that gives 6.65 if tested with a professional calibrated probe, so the reading of this cheap probe is approx. 0.3 too high, but the buffer solution is packed by the dealer, not the manufacturer. But the technical quality of the cheap probe is nothing to complain about, it's the one tap software for calibration.

  • @vernonazusa5430
    @vernonazusa54308 ай бұрын

    Great Vid! My meter hasn't been used for a while and rinsed with tap water, so now readings jump up, down, zero and keeps repeating until eventually landing in zero. Going to try your method and do a full recal to pH water used for garden vegetable seeds germinating. Ever consider sharing your methods on how you get your PPM and pH died in via diy/budget method? Would ❤ to see that! Thanks again, and happy holidays to you and your fam!

  • @docmarty809
    @docmarty8092 жыл бұрын

    Thank you It worked great. Saved 12 bucks

  • @loribadgero1035
    @loribadgero10354 ай бұрын

    Thank you for this video! My, almost, brand new meter keeps acting up after changing batteries and calibrating, twice. I tried Physan 20 to clean it 🤷🏼‍♀️, but the isopropyl alcohol and hot water trick worked. I soaked it for a short time, too, and it finally zeroed. Whew!! 😰 ❤

  • @neilrobins3329
    @neilrobins33297 ай бұрын

    Cool trick thanks. I'll try this!

  • @Karaon
    @Karaon9 ай бұрын

    worked, mine calibrated at 4.01 only and had errors on 6,86 or sth. I rinsed it nicely under hot water for a minute or two, then cleaned in distilled water. calibration went flawlessly after that, perhaps it's going to be a a routine for the cheap model if not used often

  • @brunoburtoni8479
    @brunoburtoni8479 Жыл бұрын

    Mate you really polished the shit out of that

  • @MrDabrudda

    @MrDabrudda

    Жыл бұрын

    Yep, it's shiny now

  • @burakmansurozyurt
    @burakmansurozyurt3 жыл бұрын

    thank you sir :)

  • @srmaozen4720
    @srmaozen4720 Жыл бұрын

    Hello, I need your help. I fixed the voltage to 2.50. It gives accurate results in 7.01 solution. But when you measure milk, the voltage changes. so is it normal to give different voltage in each solution?

  • @generalawareness101
    @generalawareness101 Жыл бұрын

    I had a bad EC meter that they sent a pH, and the EC, meters that are going back on Monday as one pH meter works while the brand new one would Err on me when testing. Nothing fixed it.

  • @genome692002
    @genome692002 Жыл бұрын

    thanks didnt know you have to press it again for a second time when calibrating ph 4.. the instruction manual was not very clear..

  • @Khryz-mc9nv
    @Khryz-mc9nv3 жыл бұрын

    I accidentally drop it in the solution and it's not turning on😩but this is a nice video👍

  • @trupyrodice4462

    @trupyrodice4462

    2 жыл бұрын

    Just open it up -take the batteries out and put all pieces into a bag of rice overnight, or until dry. Check to make sure it’s all dry inside and replace the batteries. You’ll be up and testing your sugar washes, fish tanks, or like, you know… your methamphetamines or whatever. Lol. I dropped mine into my sugar wash and I thought it was done for sure, but I rinsed that bad boy out and dried it up and it worked until I broke the glass bulb when taking it out of the black cap. Lol

  • @jonathanlagos9855
    @jonathanlagos98556 ай бұрын

    Gracias me salvaste una producción de cerveza.no funcionaba mí medidor de ph ahora si

  • @CookieCreamOreo
    @CookieCreamOreo3 жыл бұрын

    Great video! How long does your ph buffer solution last when you store it in the mason jars? I would love to save the money and not have to keep repurchasing it to calibrate my ph pen. I suggest you make a video on that because it would be very helpful for many. Thanks.

  • @MrDabrudda

    @MrDabrudda

    3 жыл бұрын

    I've had that PH meter for about a year. I don't cross contaminate the solutions because I alway rinse off the tip in distilled water before testing the other PH levels. I also store my solutions away from heat and light.

  • @nicoherrerai

    @nicoherrerai

    2 жыл бұрын

    Not a good idea, it changes unless you have ABSOLUTELY ZERO contamination, wich never happens in the real world

  • @barteknowakowski1962

    @barteknowakowski1962

    2 жыл бұрын

    @@nicoherrerai use distilled water, ideally :) it has ZERO contamination. And practically close to it.

  • @patriciaahadzi8024
    @patriciaahadzi80245 ай бұрын

    Please I did my calibration the first one came right , second one should came out right but the third one 9.07 instead of 9.18 and it's stuck on a particular number which is 0.78 and with I put the pH meter in my product it read number and still stop at same number 0.78 please what should I do

  • @yusritassiem9913
    @yusritassiem9913 Жыл бұрын

    it got wet on the inside so will most likely get corrosion on the inside as well which eventually is Gonna Give You a false readings or just packup

  • @michaeljames5936
    @michaeljames59362 жыл бұрын

    Your method seems to work for you, but I would caution others, that most articles I've read, say to never store the bulb in deionised water, but to use 4.01 buffer solution and that exposing it to temperatures above 50C will kibosh it.

  • @osmanfb1

    @osmanfb1

    Жыл бұрын

    He is using degrees F ! 90F = 50C

  • @andrewbissonnette6520
    @andrewbissonnette6520 Жыл бұрын

    You should do radio

  • @maddavo565
    @maddavo5654 ай бұрын

    I have a meter from Amazon that is very similar to this BUT it automatically chooses the calibration solution rather than having to press the CAL button. Pressing CAL puts it into calibration mode, but subsequent presses do nothing, it automatically decides which solution it must be in. This is BAD. It thinks my 6.86 solution is 9.18. It thinks my 4,01 solution is 6.86. I'm stuck, IDK what to do.

  • @gregandkellyone

    @gregandkellyone

    3 ай бұрын

    Same issue here, I need to reset pH calibration somehow....

  • @cimorelos

    @cimorelos

    8 күн бұрын

    @@gregandkellyone From flexiblebirdchannel: By removing the battery, it resets jammed calibration.

  • @ToluSeun
    @ToluSeun2 жыл бұрын

    how do I get mine to give accurate number? it's new. and I tried calibrating but it keeps flashing series of numbers and won't finish at the numbers on the buffers.

  • @MrDabrudda

    @MrDabrudda

    2 жыл бұрын

    You could try soaking the bulb in distilled water over night. If the bulb get dried out then it will give sporatic readings. Also check the batteries and ensure they are good.

  • @afterburner3999

    @afterburner3999

    2 жыл бұрын

    @@MrDabrudda How can Glass Bulb get Dried Out, Glass Dont Soak UP Water, if there is Water in Glass Bulb how will it GO Through glass to Refill It and What is meant to Be inside Glass Bulb water or PH7 Solution it makes no sense to try and Rehydrate Glass bcos everywhere i Look it says Glass dont hold water unless its a Glass Jar.

  • @MrDabrudda

    @MrDabrudda

    2 жыл бұрын

    "If the electrode has not been hydrated (i.e. placed in solution for more than one hour), allow the electrode to soak in a buffer (preferably pH 4) prior to standardization or measurement. This will help optimize and re-establish the thin hydration layer on the sensing bulb, which is critical to pH measurement." www.coleparmer.com/tech-article/how-to-store-clean-and-recondition-ph-electrodes

  • @hb7182
    @hb7182 Жыл бұрын

    I had bad connection battery to board. Had symbol low battery with every new one. I cannot calibrate and had Err. I resolder battery connection. After it works pH meter great.

  • @d9zirable
    @d9zirable Жыл бұрын

    7:05 my meter tries to automatically match the calibration ph, when i press the calibrate button again while it's flashing, it doesn't do anything, it will not switch. now the readings are all messed up

  • @MrDabrudda

    @MrDabrudda

    Жыл бұрын

    Try pulling out the batteries for about 10 minutes and try again.

  • @partridgeluke
    @partridgeluke2 жыл бұрын

    Mine says 1. And that's it won't change at all do you know what it is?

  • @afterburner3999

    @afterburner3999

    2 жыл бұрын

    Faulty Water.

  • @deepan02
    @deepan02 Жыл бұрын

    i no have electro ph meter

  • @knight6238
    @knight62382 жыл бұрын

    mine didnt even come with batterys

  • @afterburner3999

    @afterburner3999

    2 жыл бұрын

    Arg Ripped Off By Scamm the Battries cost More than Meter.

  • @Faris2saucey

    @Faris2saucey

    2 жыл бұрын

    Free u bro

  • @ShopperPlug
    @ShopperPlug2 жыл бұрын

    I haven't used my pH meter for about 6 years. It was covered with it's cap. I tired calibrating it with some 6 year old calibration solution... I don't know which is which is the problem lol.

  • @MrDabrudda

    @MrDabrudda

    2 жыл бұрын

    I'd soak it in warm distilled water over night and let the bulb rehydrate.

  • @MrDabrudda

    @MrDabrudda

    2 жыл бұрын

    I'd soak it in warm distilled water over night and let the bulb rehydrate.

  • @ShopperPlug

    @ShopperPlug

    2 жыл бұрын

    @@MrDabrudda Thank You. I was ready to dispose the pH meter because I kept getting errors from the old pH buffers. I will rehydrate and test again if I can get it to calibrate in some form of sensible reading. I’ll try to make my own pH solution.

  • @jaytee2456
    @jaytee24562 жыл бұрын

    funny how u edited the third blink on 6.86 cause it always show error after 2 blinks

  • @MrDabrudda

    @MrDabrudda

    2 жыл бұрын

    Right, that's why it blinked 686 because it's a conspiracy.

  • @afterburner3999
    @afterburner39992 жыл бұрын

    it might break the bulb if boilg water or wait 10 min after boiling water and will be cooler

  • @MrDabrudda

    @MrDabrudda

    2 жыл бұрын

    I NEVER said boiling water. Where in the video did I state use boiling water? Other than saying do not use boiling water.

  • @osmanfb1

    @osmanfb1

    Жыл бұрын

    @@MrDabrudda they think you are using degrees C when you say 100 degrees. You need to be clear not all viewers are from the US :-)

  • @adambeattie3032

    @adambeattie3032

    2 ай бұрын

    dude, he has his hand in the water as it's coming out of the tap. You think that's 100 celsius? This seems pretty obvious and I'm in Canada!

  • @barteknowakowski1962
    @barteknowakowski19622 жыл бұрын

    Ideally the little glass bulb should be continuously immersed in KCl solution in distilled water (4M). I wouldn't also use isopolypropylene alcohol on something that would be use in aquarium tank later on. There is no residue requiring non-polar solvents on it anyway. Use some acid solution to dissolve carbonates, that's it. PS especially using so much of isopolypropylene alcohol is NOT ecological friendly, my friend :)

  • @MrDabrudda

    @MrDabrudda

    2 жыл бұрын

    There is really no way to keep this meter's bulb submerged unless you stored it in a container of solution. I have an Arduino ph meter with a plastic screw on cap that I fill with distilled water with the pH adjusted to 4. I use there meters for making batches of liquid fertilizer for my plants and garden.

  • @generalawareness101

    @generalawareness101

    Жыл бұрын

    Incorrect, as IPA is environmentally friendly and Isopropyl alcohol is a natural product found in Mycoacia uda, Citrus hystrix, and other organisms. Not sure why you would think it wasn't ecologically friendly. Use 10k metric tonnes of it if needed.

  • @nashdrifter3629

    @nashdrifter3629

    Жыл бұрын

    hahahaha, wrong 'my friend'!

  • @notsogreat123
    @notsogreat1233 жыл бұрын

    PH solution in a canning jar with a rusty metal lid ?!?!?! Yeah that's going to work.

  • @MrDabrudda

    @MrDabrudda

    3 жыл бұрын

    And just were in this video do you see a rusty metal lid??? The canning lids have a thin plastic coating on the inside which prevent rust. NO WHERE in this video did YOU see a rusty lid. The canning jars and lids I use are brand new.

  • @afterburner3999

    @afterburner3999

    2 жыл бұрын

    @@MrDabrudda everyones a Genius

  • @rodneywaite2185
    @rodneywaite21853 жыл бұрын

    C

  • @cimabrightokafor3785

    @cimabrightokafor3785

    3 жыл бұрын

    Nice one very helpful thank you

  • @tylerfosse7336
    @tylerfosse73362 жыл бұрын

    DO NOT WHIPE WITH PAPER TOWEL!!! Do some more research. If you're going to dry it, use a cloth, and gently dab. Rubbing it can cause static to mess up the probe, these are VERY sensitive.

  • @MrDabrudda

    @MrDabrudda

    2 жыл бұрын

    A cotton cloth would probably generate more static than a paper towel anyways. Again...a $12 PH meter which is not that accurate in the first place.

  • @jameswilliford5478

    @jameswilliford5478

    Жыл бұрын

    Funny...mine came with instructions stating to dry with paper towels

  • @osmanfb1

    @osmanfb1

    Жыл бұрын

    @@jameswilliford5478 better to use filter paper.

  • @MUDSWAT
    @MUDSWAT2 жыл бұрын

    Piece of crap, don't waste your time. I have one, that did not work out of the box, I properly rinsed and calibrated too. I don't recommend this product.

  • @nicoherrerai

    @nicoherrerai

    2 жыл бұрын

    It works perfectly for me, I have to and when treated properly is VERY precise

  • @nicoherrerai

    @nicoherrerai

    2 жыл бұрын

    Maybe you got a bad one

  • @MUDSWAT

    @MUDSWAT

    2 жыл бұрын

    @@nicoherrerai Story of my life.

  • @osmanfb1

    @osmanfb1

    Жыл бұрын

    mine worked out of the box and was already calibrated. Very happy.